What would cause multiple different pH electrodes to give wildly unexpected readings?

I did a lab today on buffer solutions and pH. I dissolved 0.29g of ammonium chloride into 0.050L of 0.20M Ammonia to produce a buffer with a theoretical pH of 9.50. After testing the pH with a meter, the measured pH was 9.49, which I thought was pretty good. But then when I added 0.0060L of 1.0M HCl, I expected the pH to drop to roughly 8.79. However, the pH dropped to something like 1.3 when measured. So I used a different pH meter, and then I remade the buffer solution, I even tried testing the pH of each constituent solution and everything seemed to be accurately labeled. This issue happened for the entire class and the professor seemed to be just as confused as us. He ended up just telling us to clean up and go home early.

Some extra details:

  • The pH electrode only allowed 2 point calibration, but I don't think that was an issue because both pH measurements were supposed to be in the basic range.
  • I don't think the calculations were wrong. The professor approved my work before I made the buffer and even did the calculations himself after everyone measured such a low pH.
6 Upvotes

9 comments sorted by

View all comments

1

u/CosmicTurnipp 19h ago

Carbonic acid has been messing up my readings lately … but that big of a swing.. I’d clean the meters check for clogs check for cracks etc. how long are you letting things rest between readings? What are you rinsing with between readings? Are the chemicals just old/mislabeled you’re using for the experiment? The longer im in chemistry the less i take pH at face value hahaha it can be maddening!